Hello all,

I have a Tektronix 2465B that powers up with the following error code:

CT TEST 84 FAIL 0C

Looking at the service manual I cannot find this particular code. I figure
that I'm mis-interpreting the code somehow. Can anyone shed a bit of light
for me?

Thanks.
  #2   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 225
Default Tektronix 2465B error code interpretation


"JW" wrote in message
news
Hello all,

I have a Tektronix 2465B that powers up with the following

error code:

CT TEST 84 FAIL 0C

Looking at the service manual I cannot find this particular

code. I figure
that I'm mis-interpreting the code somehow. Can anyone shed a

bit of light
for me?

Thanks.


I do not have one but i was researching the 2465 to buy recently
and i was looking for repair ease or cronic problems in various
scopes so this is what i remember....

CT is counter/timer option on that scope.
I remember reading somewhere there is an options service manual
, i am guessing that will probably cover the options like CT in
more detail i.e. errors/repair etc.

HTH
robb


  #3   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 1,247
Default Tektronix 2465B error code interpretation

On Mon, 15 Oct 2007 05:43:22 -0400, JW wrote:

Hello all,

I have a Tektronix 2465B that powers up with the following error code:

CT TEST 84 FAIL 0C

Looking at the service manual I cannot find this particular code. I figure
that I'm mis-interpreting the code somehow. Can anyone shed a bit of light
for me?

Thanks.


That the problem is in the CT and not hte scope?
  #4   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 3,103
Default Tektronix 2465B error code interpretation

JW wrote in
news
Hello all,

I have a Tektronix 2465B that powers up with the following error code:

CT TEST 84 FAIL 0C

Looking at the service manual I cannot find this particular code. I
figure that I'm mis-interpreting the code somehow. Can anyone shed a
bit of light for me?

Thanks.


CT is counter-timer,an option that has a separate service manual.
It probably will be hard to find one.

there also was a pricey extender kit for servicing the options.
(they plug into a options board on the top of the scope chassis)
You can probably make your own extender ribbon cables.

also,there will be a sticker on the back of the scope telling you what
options are installed,if any.
there also were scope handle stickers like "2465B CTS" or 2465BDM"

--
Jim Yanik
jyanik
at
kua.net
  #5   Report Post  
Posted to sci.electronics.repair
JW JW is offline
external usenet poster
 
Posts: 519
Default Tektronix 2465B error code interpretation

On 15 Oct 2007 15:36:12 GMT, Jim Yanik wrote:

JW wrote in
news
Hello all,

I have a Tektronix 2465B that powers up with the following error code:

CT TEST 84 FAIL 0C

Looking at the service manual I cannot find this particular code. I
figure that I'm mis-interpreting the code somehow. Can anyone shed a
bit of light for me?

Thanks.


CT is counter-timer,an option that has a separate service manual.
It probably will be hard to find one.

there also was a pricey extender kit for servicing the options.
(they plug into a options board on the top of the scope chassis)
You can probably make your own extender ribbon cables.

also,there will be a sticker on the back of the scope telling you what
options are installed,if any.
there also were scope handle stickers like "2465B CTS" or 2465BDM"


Thanks very much Jim (and robb and PeterD). If I can't repair the
option, would you happen to know if it's possible to just remove it and
have the error go away?


  #6   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 225
Default Tektronix 2465B error code interpretation


"JW" wrote in message
...
On 15 Oct 2007 15:36:12 GMT, Jim Yanik

wrote:

JW wrote in
news
Hello all,

I have a Tektronix 2465B that powers up with the following

error code:

CT TEST 84 FAIL 0C

Looking at the service manual I cannot find this particular

code. I
figure that I'm mis-interpreting the code somehow. Can

anyone shed a
bit of light for me?

Thanks.


CT is counter-timer,an option that has a separate service

manual.
It probably will be hard to find one.

there also was a pricey extender kit for servicing the

options.
(they plug into a options board on the top of the scope

chassis)
You can probably make your own extender ribbon cables.

also,there will be a sticker on the back of the scope telling

you what
options are installed,if any.
there also were scope handle stickers like "2465B CTS" or

2465BDM"

Thanks very much Jim (and robb and PeterD). If I can't repair

the
option, would you happen to know if it's possible to just

remove it and
have the error go away?


i think you can by pass the error with a combination button press
, i.e. hold three certain buttons and scope bypasses but i do
not know affect on the performance and operation ?

also there is a way to do a option calibrate or reset or clear
??? using some menu but my memory fails me

i am sure someone else can help with that,
hope you fix it
robb



  #7   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 3,103
Default Tektronix 2465B error code interpretation

JW wrote in
:

On 15 Oct 2007 15:36:12 GMT, Jim Yanik wrote:

JW wrote in
news
Hello all,

I have a Tektronix 2465B that powers up with the following error code:

CT TEST 84 FAIL 0C

Looking at the service manual I cannot find this particular code. I
figure that I'm mis-interpreting the code somehow. Can anyone shed a
bit of light for me?

Thanks.


CT is counter-timer,an option that has a separate service manual.
It probably will be hard to find one.

there also was a pricey extender kit for servicing the options.
(they plug into a options board on the top of the scope chassis)
You can probably make your own extender ribbon cables.

also,there will be a sticker on the back of the scope telling you what
options are installed,if any.
there also were scope handle stickers like "2465B CTS" or 2465BDM"


Thanks very much Jim (and robb and PeterD). If I can't repair the
option, would you happen to know if it's possible to just remove it and
have the error go away?


Sorry,I can't recall,and no longer have access to the manuals.
It's not difficult to remove the CT card,so you could try it.
You may have to relocate some jumpers on the 2465 motherboard,the ones for
the signal pickoff for the CT option.

--
Jim Yanik
jyanik
at
kua.net
  #8   Report Post  
Posted to sci.electronics.repair
JW JW is offline
external usenet poster
 
Posts: 519
Default Tektronix 2465B error code interpretation

On 15 Oct 2007 20:39:23 GMT Jim Yanik wrote in Message
id: :

JW wrote in
:

On 15 Oct 2007 15:36:12 GMT, Jim Yanik wrote:

JW wrote in
news
Hello all,

I have a Tektronix 2465B that powers up with the following error code:

CT TEST 84 FAIL 0C

Looking at the service manual I cannot find this particular code. I
figure that I'm mis-interpreting the code somehow. Can anyone shed a
bit of light for me?

Thanks.

CT is counter-timer,an option that has a separate service manual.
It probably will be hard to find one.

there also was a pricey extender kit for servicing the options.
(they plug into a options board on the top of the scope chassis)
You can probably make your own extender ribbon cables.

also,there will be a sticker on the back of the scope telling you what
options are installed,if any.
there also were scope handle stickers like "2465B CTS" or 2465BDM"


Thanks very much Jim (and robb and PeterD). If I can't repair the
option, would you happen to know if it's possible to just remove it and
have the error go away?


Sorry,I can't recall,and no longer have access to the manuals.
It's not difficult to remove the CT card,so you could try it.
You may have to relocate some jumpers on the 2465 motherboard,the ones for
the signal pickoff for the CT option.


Thanks Jim, I was unsuccessful in repairing it as I cannot find any
schematics, (It would be difficult even with them, as the problem was
intermittent, and access to the board without an extender would be
impossible) but was able to remove the option by disconnecting all the
cables from the CT card (option 06) to the A1 and A5 boards, and jumpering
the following points on the A1 board:

J101 1-3 and 6-8.
J102 3-4 and 7-8.

This has solved the problem with the error code, and has put the scope
back in service.

Many thanks once again!
  #9   Report Post  
Posted to sci.electronics.repair
JW JW is offline
external usenet poster
 
Posts: 519
Default Tektronix 2465B error code interpretation

On Mon, 15 Oct 2007 16:35:49 -0400 "robb" wrote in
Message id: :


"JW" wrote in message
.. .
On 15 Oct 2007 15:36:12 GMT, Jim Yanik

wrote:

JW wrote in
news
Hello all,

I have a Tektronix 2465B that powers up with the following

error code:

CT TEST 84 FAIL 0C

Looking at the service manual I cannot find this particular

code. I
figure that I'm mis-interpreting the code somehow. Can

anyone shed a
bit of light for me?

Thanks.

CT is counter-timer,an option that has a separate service

manual.
It probably will be hard to find one.

there also was a pricey extender kit for servicing the

options.
(they plug into a options board on the top of the scope

chassis)
You can probably make your own extender ribbon cables.

also,there will be a sticker on the back of the scope telling

you what
options are installed,if any.
there also were scope handle stickers like "2465B CTS" or

2465BDM"

Thanks very much Jim (and robb and PeterD). If I can't repair

the
option, would you happen to know if it's possible to just

remove it and
have the error go away?


i think you can by pass the error with a combination button press
, i.e. hold three certain buttons and scope bypasses but i do
not know affect on the performance and operation ?


Yeah I know, but this failure for some reason was causing intermittent
sweep problems.

also there is a way to do a option calibrate or reset or clear
??? using some menu but my memory fails me


I couldn't find anything in the 2465B service manual about this. Oh
well...

i am sure someone else can help with that,
hope you fix it
robb


Thanks for your help, robb.
  #10   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 26
Default Tektronix 2465B error code interpretation

On 17 Okt., 11:34, JW wrote:
On Mon, 15 Oct 2007 16:35:49 -0400 "robb" wrote in
Message id: :







"JW" wrote in message
.. .
On 15 Oct 2007 15:36:12 GMT, Jim Yanik

wrote:


JW wrote in
news


Hello all,


I have a Tektronix 2465B that powers up with the following

error code:


CT TEST 84 FAIL 0C


Looking at the service manual I cannot find this particular

code. I
figure that I'm mis-interpreting the code somehow. Can

anyone shed a
bit of light for me?


Thanks.


CT is counter-timer,an option that has a separate service

manual.
It probably will be hard to find one.


there also was a pricey extender kit for servicing the

options.
(they plug into a options board on the top of the scope

chassis)
You can probably make your own extender ribbon cables.


also,there will be a sticker on the back of the scope telling

you what
options are installed,if any.
there also were scope handle stickers like "2465B CTS" or

2465BDM"


Thanks very much Jim (and robb and PeterD). If I can't repair

the
option, would you happen to know if it's possible to just

remove it and
have the error go away?


i think you can by pass the error with a combination button press
, i.e. hold three certain buttons and scope bypasses but i do
not know affect on the performance and operation ?


Yeah I know, but this failure for some reason was causing intermittent
sweep problems.

also there is a way to do a option calibrate or reset or clear
??? using some menu but my memory fails me


I couldn't find anything in the 2465B service manual about this. Oh
well...

i am sure someone else can help with that,
hope you fix it
robb


Thanks for your help, robb.- Zitierten Text ausblenden -

- Zitierten Text anzeigen -


Go to bama.sbc.edu (or the mirror site), there you can download a
manual with schematics, all error information, etc.



hth,
Andreas




  #11   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 3,103
Default Tektronix 2465B error code interpretation

Andreas Tekman wrote in
ps.com:

On 17 Okt., 11:34, JW wrote:
On Mon, 15 Oct 2007 16:35:49 -0400 "robb" wrote in
Message id: :







"JW" wrote in message
.. .
On 15 Oct 2007 15:36:12 GMT, Jim Yanik
wrote:


JW wrote in
news


Hello all,


I have a Tektronix 2465B that powers up with the following
error code:


CT TEST 84 FAIL 0C


Looking at the service manual I cannot find this particular
code. I
figure that I'm mis-interpreting the code somehow. Can
anyone shed a
bit of light for me?


Thanks.


CT is counter-timer,an option that has a separate service
manual.
It probably will be hard to find one.


there also was a pricey extender kit for servicing the
options.
(they plug into a options board on the top of the scope
chassis)
You can probably make your own extender ribbon cables.


also,there will be a sticker on the back of the scope telling
you what
options are installed,if any.
there also were scope handle stickers like "2465B CTS" or
2465BDM"


Thanks very much Jim (and robb and PeterD). If I can't repair
the
option, would you happen to know if it's possible to just
remove it and
have the error go away?


i think you can by pass the error with a combination button press
, i.e. hold three certain buttons and scope bypasses but i do
not know affect on the performance and operation ?


Yeah I know, but this failure for some reason was causing intermittent
sweep problems.

also there is a way to do a option calibrate or reset or clear
??? using some menu but my memory fails me


I couldn't find anything in the 2465B service manual about this. Oh
well...

i am sure someone else can help with that,
hope you fix it
robb


Thanks for your help, robb.- Zitierten Text ausblenden -

- Zitierten Text anzeigen -


Go to bama.sbc.edu (or the mirror site), there you can download a
manual with schematics, all error information, etc.



hth,
Andreas




For the OPTIONS?
The options all came with separate service manuals.( a royal PITA)
And I believe they are kinda hard to come by these days.

When I think of all the TEK manuals tossed out when the Orlando Service
center closed...I coulda been RICH,if only I had had a place to store all
of them.

--
Jim Yanik
jyanik
at
kua.net
  #12   Report Post  
Posted to sci.electronics.repair
JW JW is offline
external usenet poster
 
Posts: 519
Default Tektronix 2465B error code interpretation

On 22 Oct 2007 16:47:26 GMT Jim Yanik wrote in Message
id: :

Andreas Tekman wrote in
ups.com:

On 17 Okt., 11:34, JW wrote:


[snippo]

I couldn't find anything in the 2465B service manual about this. Oh
well...

i am sure someone else can help with that,
hope you fix it
robb

Thanks for your help, robb.- Zitierten Text ausblenden -

- Zitierten Text anzeigen -


Go to bama.sbc.edu (or the mirror site), there you can download a
manual with schematics, all error information, etc.



hth,
Andreas




For the OPTIONS?
The options all came with separate service manuals.( a royal PITA)
And I believe they are kinda hard to come by these days.


Yep. I had no luck at all finding the schematic for the CT option. At ANY
price. Oh well, without the option the scope works fine, if a bit less
capable.

When I think of all the TEK manuals tossed out when the Orlando Service
center closed...I coulda been RICH,if only I had had a place to store all
of them.


Tossed out?

*sob*
  #13   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 3,103
Default Tektronix 2465B error code interpretation

JW wrote in
:

On 22 Oct 2007 16:47:26 GMT Jim Yanik wrote in
Message id: :

Andreas Tekman wrote in
oups.com:

On 17 Okt., 11:34, JW wrote:


[snippo]

I couldn't find anything in the 2465B service manual about this. Oh
well...

i am sure someone else can help with that,
hope you fix it
robb

Thanks for your help, robb.- Zitierten Text ausblenden -

- Zitierten Text anzeigen -

Go to bama.sbc.edu (or the mirror site), there you can download a
manual with schematics, all error information, etc.



hth,
Andreas




For the OPTIONS?
The options all came with separate service manuals.( a royal PITA)
And I believe they are kinda hard to come by these days.


Yep. I had no luck at all finding the schematic for the CT option. At
ANY price. Oh well, without the option the scope works fine, if a bit
less capable.

When I think of all the TEK manuals tossed out when the Orlando
Service center closed...I coulda been RICH,if only I had had a place
to store all of them.


Tossed out?

*sob*


Yeah,and the same happened at all the other 7-8 TEK field offices that got
closed back in 1999. :-(
(there used to be THIRTY TEK field offices!!!)

PLUS,all the TEK semiconductor reference manuals [company confidential]
that got shredded/tossed! Now they put all that on the company intranet,and
only they can access it.

--
Jim Yanik
jyanik
at
kua.net
  #14   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 225
Default Tektronix 2465B error code interpretation


"JW" wrote in message
news
Hello all,

I have a Tektronix 2465B that powers up with the following

error code:

CT TEST 84 FAIL 0C

Looking at the service manual I cannot find this particular

code. I figure
that I'm mis-interpreting the code somehow. Can anyone shed a

bit of light
for me?

Thanks.


hey JW,
i hate having problems like that thus it stuck in my head
so... while searching for info on 2400 series (considering
purchase) ...

found your error

Check : CT
Test : 84
Routine: Gate Array (U6180) trigger path tests
Error : 0C
Boolean OR trigger failed to generate a sweep on
the rising edge of ATS.
Check U6180, Q5981, Q5982, Q5983, Q6090, Q6092,
Q6091, U5952, U6070

hope this helps
robb


  #15   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 3,103
Default Tektronix 2465B error code interpretation

"robb" wrote in
:


"JW" wrote in message
news
Hello all,

I have a Tektronix 2465B that powers up with the following

error code:

CT TEST 84 FAIL 0C

Looking at the service manual I cannot find this particular

code. I figure
that I'm mis-interpreting the code somehow. Can anyone shed a

bit of light
for me?

Thanks.


hey JW,
i hate having problems like that thus it stuck in my head
so... while searching for info on 2400 series (considering
purchase) ...

found your error

Check : CT
Test : 84
Routine: Gate Array (U6180) trigger path tests
Error : 0C
Boolean OR trigger failed to generate a sweep on
the rising edge of ATS.
Check U6180, Q5981, Q5982, Q5983, Q6090, Q6092,
Q6091, U5952, U6070

hope this helps
robb




if you could scan and upload your CT option manual to BAMA,it would be
greatly appreciated. Same for any of the other 2445/65 options.

--
Jim Yanik
jyanik
at
kua.net
Reply
Thread Tools Search this Thread
Search this Thread:

Advanced Search
Display Modes

Posting Rules

Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On


Similar Threads
Thread Thread Starter Forum Replies Last Post
tektronix 2465 error code JMINN699 Electronics Repair 1 April 5th 07 06:15 PM
Tektronix 2467B (similar to 2465B): Error Message TEST 04 FAIL 13 (3453) JW Electronics Repair 4 January 9th 07 07:14 AM
Tektronix 2467B (similar to 2465B): Error Message TEST 04 FAIL 13 (3453) tek_user Electronics Repair 4 December 2nd 06 07:05 PM
Tektronix 2465B Problems Peter Electronics Repair 3 October 16th 04 05:14 AM


All times are GMT +1. The time now is 01:08 AM.

Powered by vBulletin® Copyright ©2000 - 2024, Jelsoft Enterprises Ltd.
Copyright ©2004-2024 DIYbanter.
The comments are property of their posters.
 

About Us

"It's about DIY & home improvement"