Default tek 545 scope 'too bright'

Have had this 545 sitting, unused, for years, in my utility room. Bought it
surplus for 50 bux about 20 years ago. IIRC, it 'used to work fine'.

I wheeled it into my ham shack last week, and turned it on. The trace was
too bright and would not turn down completely. After a few minutes, it shut
down. Found that the FAN mounts had deteriorated and the cooling fan was
not turning. After making new fan mounts and digging lots of small pieces
of aluminum (from the fan hitting the filter) out of the scope, I turned it
on again. Still too bright. Focus/astigmatism are poor also.

Beam blanking seems to be working as the spot goes away when there is no
sweep.

Checking voltages, the 220 volt test point on the upper right side is only
about 135. Other voltages (500, 350, 100, -150) are all good.

Is anyone familiar enought with this old workhorse of a scope so that they
can tell me which components are most likely to cause this problem?

Also, I would expect that there would be some key or pictures of the layout
of the beast that would show me exactly WHERE each part is located. For
example, I have a capacitor (black body) on the right side, in what looks
like the delayed sweep section, that has a crack along the side of it. I
cut one lead and measured. Doesn't seem to be leaky. Can not read the color
code on the part. I would like to know which part it is so can compare a
cap meter reading with the spec for the part WITHOUT having to trace wires
and figure out which cap it is. Does anyone know if there is a simple
correlation between part locations and names or part numbers or a parts
layout drawing available?

Checked the tubes in the power supply section and found 1/2 of a 12AX7 that
is marginal. My box of tubes that used to reside in my carport has
disappeared, some years ago. Either stolen, given away or discarded while
cleaning up, so I have ordered a russian knockoff from a tube supply house.
I did try a 12AT7 from an extra plug in, but that didn't fix the bright
trace.



--
bz

please pardon my infinite ignorance, the set-of-things-I-do-not-know is an
infinite set.




--
bz 73 de N5BZ k

please pardon my infinite ignorance, the set-of-things-I-do-not-know is an
infinite set.

remove ch100-5 to avoid spam trap
  #2   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 129
Default tek 545 scope 'too bright'

if you go download:

ftp://bama.sbc.edu/downloads/tek/545...%20545a(2).pdf

you'll have a service manual.

You probably should replace any of those black tubular bakelite
capacitors.. They tend to get anywhere from slightly to horribly
leaky. Any leakage over a microamp will mess up the grid biasing.


see the next to last page for the CRT circuit.

The brightness circuit has many high-megohm resistors-- thesetend to
drift in value over the decades.

Also around the intensity control there are several capacitors in the
C82x range -- replace them. They might measure okay under 1.5 volts
from your meter, but leak badly when under high voltage.

  #3   Report Post  
Posted to sci.electronics.repair
bz bz is offline
external usenet poster
 
Posts: 314
Default tek 545 scope 'too bright'

"Ancient_Hacker" wrote in news:1169569646.335974.73670
@d71g2000cwa.googlegroups.com:

if you go download:

ftp://bama.sbc.edu/downloads/tek/545...%20545a(2).pdf

you'll have a service manual.

You probably should replace any of those black tubular bakelite
capacitors.. They tend to get anywhere from slightly to horribly
leaky. Any leakage over a microamp will mess up the grid biasing.


That would explain a lot.

see the next to last page for the CRT circuit.

The brightness circuit has many high-megohm resistors-- thesetend to
drift in value over the decades.

Also around the intensity control there are several capacitors in the
C82x range -- replace them. They might measure okay under 1.5 volts
from your meter, but leak badly when under high voltage.


Thanks, will give those a try.

The russian 12ax7s arrived today but did not fix the problem.
Looks like it is going to be a long process to fix up this beauty.




--
bz 73 de N5BZ k

please pardon my infinite ignorance, the set-of-things-I-do-not-know is an
infinite set.

remove ch100-5 to avoid spam trap
  #4   Report Post  
Posted to sci.electronics.repair
Bob Bob is offline
external usenet poster
 
Posts: 3
Default tek 545 scope 'too bright'


"bz" wrote in message
98.139...
Have had this 545 sitting, unused, for years, in my utility room. Bought
it
surplus for 50 bux about 20 years ago. IIRC, it 'used to work fine'.

I wheeled it into my ham shack last week, and turned it on. The trace was
too bright and would not turn down completely. After a few minutes, it
shut
down. Found that the FAN mounts had deteriorated and the cooling fan was
not turning. After making new fan mounts and digging lots of small pieces
of aluminum (from the fan hitting the filter) out of the scope, I turned
it
on again. Still too bright. Focus/astigmatism are poor also.

Beam blanking seems to be working as the spot goes away when there is no
sweep.

Checking voltages, the 220 volt test point on the upper right side is only
about 135. Other voltages (500, 350, 100, -150) are all good.



There's your starting point - if the power supplies are way off that can
cause a lot of problems. Get your 220 supply back within specs and see how
the scope works. From what I remember, the -150 is the reference supply
that the rest are set to.

Granted, there could be something elsewhere in the scope loading down the
220V power supply (leaky caps, etc.) but it's the place to start
troubleshooting.

Bob



  #5   Report Post  
Posted to sci.electronics.repair
bz bz is offline
external usenet poster
 
Posts: 314
Default tek 545 scope 'too bright'

"Bob" wrote in
:


"bz" wrote in message
98.139...
Have had this 545 sitting, unused, for years, in my utility room.
Bought it
surplus for 50 bux about 20 years ago. IIRC, it 'used to work fine'.

I wheeled it into my ham shack last week, and turned it on. The trace
was too bright and would not turn down completely. After a few minutes,
it shut
down. Found that the FAN mounts had deteriorated and the cooling fan
was not turning. After making new fan mounts and digging lots of small
pieces of aluminum (from the fan hitting the filter) out of the scope,
I turned it
on again. Still too bright. Focus/astigmatism are poor also.

Beam blanking seems to be working as the spot goes away when there is
no sweep.

Checking voltages, the 220 volt test point on the upper right side is
only about 135. Other voltages (500, 350, 100, -150) are all good.



There's your starting point - if the power supplies are way off that can
cause a lot of problems. Get your 220 supply back within specs and see
how the scope works. From what I remember, the -150 is the reference
supply that the rest are set to.

Granted, there could be something elsewhere in the scope loading down
the 220V power supply (leaky caps, etc.) but it's the place to start
troubleshooting.



Smurfies law.... I pulled tubes and tested them. When I put them back in, I
couldn't even get the -150 to adjust at all. Maybe I should not have
'straightened the pins' on the tubes. Maybe I have some sockets that are
not making good contact with straight pins.

The russian 12ax7 doesn't help either.

Now, the -150 is about -87 volts and only a very small adjustment range.
The 100 volts is about 130

Looking at resistances, I am seeing a 100 ohm 'load' to ground on the 100
volt line. That CAN NOT be right. That would require a 100 ohm 100 watt
resistor and there is no such animal.

I was afraid of the bright beam damaging the CRT, even when 'off screen' so
I tried removing the CRT's socket. BAD idea. No load on the HV supply
allows the voltage to go very high and the scope protests with pops and
crackles.

Current plans are to track down the 100 ohm 'short'(or find something that
justifies it),


Bob









--
bz 73 de N5BZ k

please pardon my infinite ignorance, the set-of-things-I-do-not-know is an
infinite set.

remove ch100-5 to avoid spam trap


  #6   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 129
Default tek 545 scope 'too bright'



On Jan 25, 8:08 am, bz wrote:


Looking at resistances, I am seeing a 100 ohm 'load' to ground on the 100
volt line. That CAN NOT be right. That would require a 100 ohm 100 watt
resistor and there is no such animal.



Well, you get brownie points by using your ohmmeter to find problems,
rather than just plugging it in.

*But* in many Tek scopes they put eight 12-volt tube heaters in series
and drive them from the 100 volt line. Mainly for the tubes in the
low-signal sections, to avoid AC heater hum getting into the circuits.

And tube heaters have very low resistaance when cold-- when hot they'll
draw 150 milliamps from the 100 volt line -- much closer to reality.

I'd try scoping the plates of the voltage regulator tubes-- often the
electrolytics dry up and you end up with low and hummy regulated
voltages.

You might try switching the 6080 tubes-- you might have the weaker one
now on the 100 volt line.









I was afraid of the bright beam damaging the CRT, even when 'off screen' so
I tried removing the CRT's socket. BAD idea. No load on the HV supply
allows the voltage to go very high and the scope protests with pops and
crackles.

Current plans are to track down the 100 ohm 'short'(or find something that
justifies it),



Bob--

bz 73 de N5BZ k

please pardon my infinite ignorance, the set-of-things-I-do-not-know is an
infinite set.

remove ch100-5 to avoid spam trap


  #7   Report Post  
Posted to sci.electronics.repair
bz bz is offline
external usenet poster
 
Posts: 314
Default tek 545 scope 'too bright'

"Ancient_Hacker" wrote in
oups.com:



On Jan 25, 8:08 am, bz wrote:


Looking at resistances, I am seeing a 100 ohm 'load' to ground on the
100 volt line. That CAN NOT be right. That would require a 100 ohm 100
watt resistor and there is no such animal.



Well, you get brownie points by using your ohmmeter to find problems,
rather than just plugging it in.


Brownie points don't fix things.

*But* in many Tek scopes they put eight 12-volt tube heaters in series
and drive them from the 100 volt line. Mainly for the tubes in the
low-signal sections, to avoid AC heater hum getting into the circuits.


AHHHHHHHHHH. And I kept skipping over the filament wiring schematics,
thinking to myself 'these can't have anything to do with the problem'.

Hm.... I see TWO 12 volt tubes, V180 and V130(I think), a 12AL5 and a 12AU6
with filaments in series, off of the +100 volt line, going to a point labled
+75 to pin 15 interconnecting socket vertical amplifier... now to check the
schematic for the vertical amp...

What-do-you-know... 7 more tube filaments in the same string, hiding in the
plug in module. Well, I guess I can pull the module and see if my 100 ohms
goes up to 2k or so and then I will know that you were 'spot on'.

And tube heaters have very low resistaance when cold-- when hot they'll
draw 150 milliamps from the 100 volt line -- much closer to reality.

I'd try scoping the plates of the voltage regulator tubes-- often the
electrolytics dry up and you end up with low and hummy regulated
voltages.


Waiting for the Capacitor ESR meter kit I ordered to arrive. It looks like a
cool instrument to have around the shack!


You might try switching the 6080 tubes-- you might have the weaker one
now on the 100 volt line.


And I might have mixed them up while testing them.

Also considering replacing the high voltage rectifiers with solid state
diodes. Wonder if removing the filament load from the supply will mess things
up. Probably!

THANKS again. Experience! It makes a difference.

-bz-



I was afraid of the bright beam damaging the CRT, even when 'off
screen' so I tried removing the CRT's socket. BAD idea. No load on the
HV supply allows the voltage to go very high and the scope protests
with pops and crackles.

Current plans are to track down the 100 ohm 'short'(or find something
that justifies it),





--
bz 73 de N5BZ k

please pardon my infinite ignorance, the set-of-things-I-do-not-know is an
infinite set.

remove ch100-5 to avoid spam trap
  #8   Report Post  
Posted to sci.electronics.repair
bz bz is offline
external usenet poster
 
Posts: 314
Default tek 545 scope 'too bright'

The saga continues....

Removing the plug in module DID clear the 100 ohm 'short' to ground on the
100 volt line.

Swapping tubes did NOT fix the -150 volt supply however.
It only measures -131 volts (after several minutes of 'post hv turn on'
warm up). It drifts much more during the warm up than the -150 volt
adjustment pot will swing to voltage. In fact I see very little change as I
adjust the pot.

The 100 volt supply is putting out about 130 volts and the rest of the
supplies look about normal (have not checked the HT supply to the CRT. I
wonder if it being too high could cause all my symptoms).

Measured the current out of the -150 and 100 volt supplies (across the 5
and 10 ohm series resistors). ~.95 volts /5 ohms = 190 ma for the -150
supply
~1.4 volts/10 ohm = 140 ma for the 100 volt supply. I am assuming that
those values are not unusual. I have not looked at ripple, etc., yet.

Looks like I am going to have to measure voltages at several points and try
to make sense of any patterns that emerge.






--
bz 73 de N5BZ k

please pardon my infinite ignorance, the set-of-things-I-do-not-know is an
infinite set.

remove ch100-5 to avoid spam trap
  #9   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 3,103
Default tek 545 scope 'too bright'

bz wrote in
98.139:

The saga continues....

Removing the plug in module DID clear the 100 ohm 'short' to ground on
the 100 volt line.

Swapping tubes did NOT fix the -150 volt supply however.
It only measures -131 volts (after several minutes of 'post hv turn
on' warm up). It drifts much more during the warm up than the -150
volt adjustment pot will swing to voltage. In fact I see very little
change as I adjust the pot.

The 100 volt supply is putting out about 130 volts and the rest of the
supplies look about normal (have not checked the HT supply to the CRT.
I wonder if it being too high could cause all my symptoms).

Measured the current out of the -150 and 100 volt supplies (across the
5 and 10 ohm series resistors). ~.95 volts /5 ohms = 190 ma for the
-150 supply
~1.4 volts/10 ohm = 140 ma for the 100 volt supply. I am assuming that
those values are not unusual. I have not looked at ripple, etc., yet.

Looks like I am going to have to measure voltages at several points
and try to make sense of any patterns that emerge.







in troubleshooting those old tube scopes,I used to keep a few electrolytic
caps with clip leads to bridge the PS electrolytics,when I had a low
supply.

--
Jim Yanik
jyanik
at
kua.net
  #10   Report Post  
Posted to sci.electronics.repair
bz bz is offline
external usenet poster
 
Posts: 15
Default tek 545 scope 'too bright'

Jim Yanik wrote in
:

bz wrote in
98.139:

The saga continues....

.....


in troubleshooting those old tube scopes,I used to keep a few
electrolytic caps with clip leads to bridge the PS electrolytics,when I
had a low supply.


Good idea. Thanks!
(note to self: remember to discharge any such caps as they can 'build up'
a charge again, once disconnected.)



--
bz

please pardon my infinite ignorance, the set-of-things-I-do-not-know is an
infinite set.




  #11   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 272
Default tek 545 scope 'too bright'

"bz" wrote in message
98.139...
The saga continues....

Removing the plug in module DID clear the 100 ohm 'short' to ground on the
100 volt line.

Swapping tubes did NOT fix the -150 volt supply however.
It only measures -131 volts (after several minutes of 'post hv turn on'
warm up). It drifts much more during the warm up than the -150 volt
adjustment pot will swing to voltage. In fact I see very little change as I
adjust the pot.

The 100 volt supply is putting out about 130 volts and the rest of the
supplies look about normal (have not checked the HT supply to the CRT. I
wonder if it being too high could cause all my symptoms).

Measured the current out of the -150 and 100 volt supplies (across the 5
and 10 ohm series resistors). ~.95 volts /5 ohms = 190 ma for the -150
supply
~1.4 volts/10 ohm = 140 ma for the 100 volt supply. I am assuming that
those values are not unusual. I have not looked at ripple, etc., yet.

Looks like I am going to have to measure voltages at several points and try
to make sense of any patterns that emerge.

--
bz 73 de N5BZ k



I don't remember whether you posted that your scope is the older 545 model, or
the later 545A or 545B model.
I still have a copy of the 545B scope handy. As I remember, the low voltage
power supplies are pretty close from the basic 545 thru to the 545B models.
According to my manual, the resistances to ground (power off and all supplies
totally discharged, no plug-in installed) a
-150V 2K
+100V 2K
+225V 2K
+350V 1K
+500V 15K

Ripple specs are pretty tight, and must be to specs for a good trace and
waveform display.
-150V 5mv
+100V 10mv
+225V 5mv
+350V 20mv
+500V 20mv

Going back to your earlier posts, you started with a good -150V source, but all
others were way out of tolerance ????!!!!! Kinda hard to do with this
instrument, since all the other LV sources are referenced to the -150V source.
Are you sure that your test instrument is good?? If you have another source of
known voltage, check it and eliminate or identify that as a source of error.

Next, remove and reseat all the tubes in the power supply section. Old tubes
and sockets get a bit of fungus or mold in the works, and can create all sorts
of havoc. Work the tubes in and out of the sockets several times; you might
even clean the tube pins with a bit of fine emery cloth.

Cheers!!!!!
--
Dave M
MasonDG44 at comcast dot net (Just substitute the appropriate characters in the
address)

Some days you're the dog, some days the hydrant.


  #12   Report Post  
Posted to sci.electronics.repair
bz bz is offline
external usenet poster
 
Posts: 314
Default tek 545 scope 'too bright'

"DaveM" wrote in
:

"bz" wrote in message
98.139...
The saga continues....

Removing the plug in module DID clear the 100 ohm 'short' to ground on
the 100 volt line.

Swapping tubes did NOT fix the -150 volt supply however.
It only measures -131 volts (after several minutes of 'post hv turn on'
warm up). It drifts much more during the warm up than the -150 volt
adjustment pot will swing to voltage. In fact I see very little change
as I adjust the pot.

The 100 volt supply is putting out about 130 volts and the rest of the
supplies look about normal (have not checked the HT supply to the CRT.
I wonder if it being too high could cause all my symptoms).

Measured the current out of the -150 and 100 volt supplies (across the
5 and 10 ohm series resistors). ~.95 volts /5 ohms = 190 ma for the
-150 supply
~1.4 volts/10 ohm = 140 ma for the 100 volt supply. I am assuming that
those values are not unusual. I have not looked at ripple, etc., yet.

Looks like I am going to have to measure voltages at several points and
try to make sense of any patterns that emerge.

--
bz 73 de N5BZ k



I don't remember whether you posted that your scope is the older 545
model, or the later 545A or 545B model.


It is a 545. No A or B.

I still have a copy of the 545B scope handy. As I remember, the low
voltage power supplies are pretty close from the basic 545 thru to the
545B models. According to my manual, the resistances to ground (power
off and all supplies totally discharged, no plug-in installed) a


That 'no plug-in installed' is more important than I knew!

-150V 2K
+100V 2K
+225V 2K
+350V 1K
+500V 15K


Will double check these, but I think that seems like about what I was
seeing.


Ripple specs are pretty tight, and must be to specs for a good trace and
waveform display.
-150V 5mv
+100V 10mv
+225V 5mv
+350V 20mv
+500V 20mv


Will check.


Going back to your earlier posts, you started with a good -150V source,
but all others were way out of tolerance ????!!!!!


-150 was ok as were MOST of the others:
"Checking voltages, the 220 volt test point on the upper right side is only
about 135. Other voltages (500, 350, 100, -150) are all good."

Kinda hard to do
with this instrument, since all the other LV sources are referenced to
the -150V source. Are you sure that your test instrument is good??


Moderately sure but I have several meters that I can check it with.

If
you have another source of known voltage, check it and eliminate or
identify that as a source of error.


Will do.


Next, remove and reseat all the tubes in the power supply section. Old
tubes and sockets get a bit of fungus or mold in the works, and can
create all sorts of havoc.


Oxidation, etc., can cause problems, for sure!

Work the tubes in and out of the sockets
several times; you might even clean the tube pins with a bit of fine
emery cloth.


I may do that and run a small rat tail file in and out of each pin in the
tub sockets.

I did pull the power supply tubes and check them in my tube checker.
Straightened the pins while doing that (always do in order to protect the
tub sockets in the checker). It was after I checked the tubes that the -150
volts went to hexxxx went crazy.


Cheers!!!!!


Thanks for the suggestions. Once I fix the scope, then I have another piece
of equipment that I wanted to use the scope on.






--
bz 73 de N5BZ k

please pardon my infinite ignorance, the set-of-things-I-do-not-know is an
infinite set.

remove ch100-5 to avoid spam trap
  #13   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 3,103
Default tek 545 scope 'too bright'

bz wrote in
98.139:

Jim Yanik wrote in
:

bz wrote in
98.139:

The saga continues....

....


in troubleshooting those old tube scopes,I used to keep a few
electrolytic caps with clip leads to bridge the PS electrolytics,when I
had a low supply.


Good idea. Thanks!
(note to self: remember to discharge any such caps as they can 'build up'
a charge again, once disconnected.)




I used a 1Kohm 2W resistor glued to a plastic stick to discharge caps.
Bent the leads to fit the cap's terminals.
I also used this for switcher PS line caps.

--
Jim Yanik
jyanik
at
kua.net
  #14   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 1,001
Default tek 545 scope 'too bright'

bz wrote:

The saga continues....

Removing the plug in module DID clear the 100 ohm 'short' to ground on the
100 volt line.

Swapping tubes did NOT fix the -150 volt supply however.
It only measures -131 volts (after several minutes of 'post hv turn on'
warm up). It drifts much more during the warm up than the -150 volt
adjustment pot will swing to voltage. In fact I see very little change as I
adjust the pot.

The 100 volt supply is putting out about 130 volts and the rest of the
supplies look about normal (have not checked the HT supply to the CRT. I
wonder if it being too high could cause all my symptoms).

Measured the current out of the -150 and 100 volt supplies (across the 5
and 10 ohm series resistors). ~.95 volts /5 ohms = 190 ma for the -150
supply
~1.4 volts/10 ohm = 140 ma for the 100 volt supply. I am assuming that
those values are not unusual. I have not looked at ripple, etc., yet.

Looks like I am going to have to measure voltages at several points and try
to make sense of any patterns that emerge.






and to think, i through out years ago, 2 perfectly working 545 boat
anchors.


--
"I'm never wrong, once i thought i was, but was mistaken"
Real Programmers Do things like this.
http://webpages.charter.net/jamie_5

  #15   Report Post  
Posted to sci.electronics.repair
bz bz is offline
external usenet poster
 
Posts: 314
Default tek 545 scope 'too bright'

Jamie t wrote in
:

bz wrote:

The saga continues....

Removing the plug in module DID clear the 100 ohm 'short' to ground on
the 100 volt line.

Swapping tubes did NOT fix the -150 volt supply however.
It only measures -131 volts (after several minutes of 'post hv turn on'
warm up). It drifts much more during the warm up than the -150 volt
adjustment pot will swing to voltage. In fact I see very little change
as I adjust the pot.

The 100 volt supply is putting out about 130 volts and the rest of the
supplies look about normal (have not checked the HT supply to the CRT.
I wonder if it being too high could cause all my symptoms).

Measured the current out of the -150 and 100 volt supplies (across the
5 and 10 ohm series resistors). ~.95 volts /5 ohms = 190 ma for the
-150 supply
~1.4 volts/10 ohm = 140 ma for the 100 volt supply. I am assuming that
those values are not unusual. I have not looked at ripple, etc., yet.

Looks like I am going to have to measure voltages at several points and
try to make sense of any patterns that emerge.






and to think, i through out years ago, 2 perfectly working 545 boat
anchors.


I _had_ two(with carts!). My wife "disposed" of one.

If you look at the prices on 12AX7s on the web, you will realize that you
may have thrown out thousands of dollars worth of tubes. Some vintages of
12ax7 go (to audio buffs) for unimaginable prices! Of course the russian
'knock offs' go for less than 10 bux.











--
bz 73 de N5BZ k

please pardon my infinite ignorance, the set-of-things-I-do-not-know is an
infinite set.

remove ch100-5 to avoid spam trap


  #16   Report Post  
Posted to sci.electronics.repair
external usenet poster
 
Posts: 38
Default tek 545 scope 'too bright'

bz wrote:

I may do that and run a small rat tail file in and out of each pin in the
tub sockets.


NO NO NO! The sockets contacts are plated (tin?). A file will remove the
plating, opening up the base metal to further corrosion. Instead, spray
some contact cleaner into the sockets, and then run something scrubby
but non-abrasive through them, like a pipe cleaner.

Someone else suggested touching up the tube pins with fine emery cloth.
That's OK - tubes are replaceable. But in practical terms, sockets are
not.

Bill
  #17   Report Post  
Posted to sci.electronics.repair
bz bz is offline
external usenet poster
 
Posts: 314
Default tek 545 scope 'too bright'

Bill Jeffrey wrote in
:

bz wrote:

I may do that and run a small rat tail file in and out of each pin in
the tub sockets.


NO NO NO! The sockets contacts are plated (tin?).


Sometimes tin, sometimes silver, gold on the best, IF I remember
correctly.

A file will remove the
plating, opening up the base metal to further corrosion. Instead, spray
some contact cleaner into the sockets, and then run something scrubby
but non-abrasive through them, like a pipe cleaner.


Pipe cleaners (those I am familiar with) are usually too thick for
miniature tube sockets (might work on octal but not loctal) and many are
built on a wire base that could distort the shape of the spring. Perhaps a
wooden toothpick, used with caution.


Someone else suggested touching up the tube pins with fine emery cloth.
That's OK - tubes are replaceable. But in practical terms, sockets are
not.


You have a good point. I wouldn't have taken that step unless things were
really bad. I may need to use a sharply pointed tool to 'tighten' some of
the sockets, but, again, that is a last resort.

I do think that many of the strange problems I have been seeing are due to
bad connections on the tubes. I 'made the mistake' of seating them fully.

Looking at the tube pins closely, I can see that they have been a fraction
of an inch ABOVE the 'fully seated' position for many years and the pins
have oxidized on the exposed areas. Seating them fully was thus a mistake
on my part.



Bill






--
bz 73 de N5BZ k

please pardon my infinite ignorance, the set-of-things-I-do-not-know is an
infinite set.

remove ch100-5 to avoid spam trap
Reply
Thread Tools Search this Thread
Search this Thread:

Advanced Search
Display Modes

Posting Rules

Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On


Similar Threads
Thread Thread Starter Forum Replies Last Post
FA: *New* Tek 465B scope Ancient_Hacker Electronics Repair 11 August 27th 05 06:06 AM
Need help fixing a Tek scope Juerg Electronics Repair 20 April 14th 05 12:37 AM
TEK 475 Scope-stopped Dave_s Electronics Repair 2 April 5th 05 08:38 AM
Tek 422 scope "whine" mc Electronics Repair 1 March 7th 05 01:01 PM
tek 465b scope problem Aaron Kentucky Electronics Repair 10 March 10th 04 07:30 PM


All times are GMT +1. The time now is 11:28 AM.

Powered by vBulletin® Copyright ©2000 - 2024, Jelsoft Enterprises Ltd.
Copyright ©2004-2024 DIYbanter.
The comments are property of their posters.
 

About Us

"It's about DIY & home improvement"