Electronics Repair (sci.electronics.repair) Discussion of repairing electronic equipment. Topics include requests for assistance, where to obtain servicing information and parts, techniques for diagnosis and repair, and annecdotes about success, failures and problems.

Reply
 
LinkBack Thread Tools Search this Thread Display Modes
  #1   Report Post  
Michael J. Linden, N9BDF
 
Posts: n/a
Default Tek 2236 Scope: Counter/Timer/Multimeter (CTM) Problem

I've got a problem with the Counter/Timer/Multimeter (CTM) function
of a Tektronix 2236 oscilloscope. The CTM is not operating correctly
when I try to measure the voltage on channel 1. When I try to measure
the DC voltage of channel 1 (DC Coupling), the fluorescent display
reads "DC OUCH" indicating an overvoltage EVEN when NO voltage is
being applied. When I try to measure the AC voltage of channel 1 (AC
Coupling) the fluorescent display reads "AC 0.0013" (or somewhere
around there) regardless of what voltage is applied to channel 1.

The CTM measures voltage correctly when I use the multimeter
connector inputs on the right side of the oscilloscope. Therefore, it
appears that CTM is basically functioning, but there is some sort of
problem in the coupling of the channel 1 signal into the CTM.

Anybody familiar with this problem?

Thanks, Michael

  #2   Report Post  
tekman
 
Posts: n/a
Default

Are you shure it is not a problem of setting up and using the scope?
May be a look in the operating manual might help to set things right.

You can download one for free from www.teknetelectronics.com


hth,
Andreas

  #3   Report Post  
Marcelo David
 
Posts: n/a
Default

Not connecting anything to channel 1, doesn't mean voltage = 0V, it is
"in the air" so, any voltage may be there. Try shortcircuiting the
channel, and see if the same problem appears...
  #4   Report Post  
Michael J. Linden, N9BDF
 
Posts: n/a
Default

Yes, I have the manual and I'm sure the scope is configured correctly
to read voltage off of channel 1.

Thanks, Michael


In article .com,
says...


Are you shure it is not a problem of setting up and using the scope?
May be a look in the operating manual might help to set things right.

You can download one for free from
www.teknetelectronics.com


hth,
Andreas


  #5   Report Post  
Michael J. Linden, N9BDF
 
Posts: n/a
Default

I've tried doing this with a 1 KHz 3vpp signal (AC coupling), a 1.5v
alkaline battery (DC coupling), and an open circuit (AC and DC
couplines). The behavior is the same in all three cases. I will try
short-circuiting the probe to see if the behavior changes.

Thanks, Michael


In article ,
says...


Not connecting anything to channel 1, doesn't mean voltage = 0V, it is
"in the air" so, any voltage may be there. Try shortcircuiting the
channel, and see if the same problem appears...




  #6   Report Post  
Jim Yanik
 
Posts: n/a
Default

"Michael J. Linden, N9BDF" wrote in
:

Yes, I have the manual and I'm sure the scope is configured correctly
to read voltage off of channel 1.




The proper way to proceed is to look at the schematics and circuit
description in your service manual,and check the voltage/signal that's
picked off from the Ch.1 vertical and applied to the DMM input.

--
Jim Yanik
jyanik
at
kua.net
  #7   Report Post  
Michael J. Linden, N9BDF
 
Posts: n/a
Default

OK, guys, I found the problem. The cable that routes the channel 1 signal
from the "attenuator box" had disconnected itself. It looks like the weight
of the attached toroid was probably responsible.

Thanks, Michael


In article , .
says...


"Michael J. Linden, N9BDF" wrote in
:

Yes, I have the manual and I'm sure the scope is configured correctly
to read voltage off of channel 1.




The proper way to proceed is to look at the schematics and circuit
description in your service manual,and check the voltage/signal that's
picked off from the Ch.1 vertical and applied to the DMM input.

--
Jim Yanik
jyanik
at
kua.net


Reply
Thread Tools Search this Thread
Search this Thread:

Advanced Search
Display Modes

Posting Rules

Smilies are On
[IMG] code is On
HTML code is Off
Trackbacks are On
Pingbacks are On
Refbacks are On


Similar Threads
Thread Thread Starter Forum Replies Last Post
Problem with Tektronix scope 7313 Jim Yanik Electronics Repair 4 August 5th 04 12:37 PM
Tektronix 466 Scope Problem Chris Electronics Repair 13 December 24th 03 03:50 PM
Scope advice needed Jeff Electronics Repair 10 November 18th 03 01:47 AM
BK Precision 1522 scope: bad input section, no trace Ben Byer Electronics Repair 3 September 22nd 03 10:29 PM
opinions on buying a scope tempus fugit Electronics Repair 18 July 31st 03 02:29 PM


All times are GMT +1. The time now is 07:55 PM.

Powered by vBulletin® Copyright ©2000 - 2024, Jelsoft Enterprises Ltd.
Copyright ©2004-2024 DIYbanter.
The comments are property of their posters.
 

About Us

"It's about DIY & home improvement"