Posted to alt.engineering.electrical,sci.electronics.design,sci.electronics.equipment,sci.electronics.repair
Phil Allison[_2_] Phil Allison[_2_] is offline
external usenet poster
 
Posts: 1,716
Default Clamp meters: Peak vs. In-rush?


"daestrom"

Reading the app note it seems that Fluke designed there 33 meters to read
the symetrical currents for motor starting.

But as you mentioned, large inductive loads often have a DC offset
component to their starting current. This comes as an artifact of closing
the starter when the sine wave is not at zero-crossing


** Fraid you have got that all wrong.

The way to *guarantee* very large inrush surges ( with transformers and
transformer based PSUs) is to switch on at the zero crossing of the AC
voltage.

Cos doing this generates the maximum degree of magnetic saturation in the
core.


I wonder if Fluke deliberately filter out the DC offset just so they don't
have to explain why the reading changes each time you start the motor :-)



** Fluke make no specific claims about the accuracy of their "inrush surge"
detection circuitry.

But I would not doubt is does the job required, as far as motors and
circuit breakers are concerned.


..... Phil