Joseph Gwinn wrote:
In article ,
Spehro Pefhany wrote:

On Sat, 04 Apr 2009 19:50:29 -0400, the renowned Joseph Gwinn
wrote:


As for manufacture by a HSM, either approach is practical, but the Wyler
approach is far simpler mechanically. In either case, the electronics
part is dead simple by present-day standards, but was a big deal back in
the 1960s and 1970s.


Joe Gwinn

Have you considered the liquid type? They claim 5 arc second
repeatability.


Yes, I did consider electrolytric types (a bubble level vial with
electronic readout). The pendulum units are a factor of ten better in
resolution and repeatability, 0.1 or 0.2 arc seconds versus 5 arc
seconds. Long term drift is also an issue. It sure would be easier if
the electrolytic sensors were better than they are.

Believe me, .1 arc seconds is AWFULLY sensitive.
I can walk from one end of my Sheldon lathe to the
other and tilt it a full arc second by deflecting
the concrete floor.
So, unless you are aligning turbo-alternator sets
in nuclear power plants, you have no need for such
sensitivity. Possibly because my Talyvel is old
and hasn't been to the factory in 20+ years, it
also drifts. If I am going to be doing anything
precise, I turn it on and let it "warm up" for
5-10 minutes while I am setting up. Then it is
pretty stable. It may still drift a full arc
second or so per hour, may be temperature changes
as I handle it a lot when checking surfaces.