Need help fixing a Tek scope
Hi,
I could use some help fixing a TDS640 digital Tek scope. The baselines of all four channels show an offset'ed sawtooth (different amplitudes and offsets for each channel) with a period of 4.5ms rather than a flat 0. When applying an input signal, it gets overlaid over the sawtooth but otherwise looks fine. I was following Teks troubleshooting guide and located the problem to the acquisition board. The input attenuator is believed to be OK (input to the AD converters look fine). All the low voltages (+/-15V, +/-5V) are OK and within spec (checked with another scope, no ripples). What I noticed is that over time when the scope warms up the amplitude of the sawtooth decreases until it becomes zero, but there's still an offset that won't go away. I was using some freeze spray to try to isolate the problem further and ended up pin-pointing it to the section around the AD converters. When I cold spray that area, the sawtooth comes back and disappears again after a while. There's a bunch of OpAmps, resisors, caps and diodes in that area. Any ideas what to look at first? Thanks in advance for any suggestions ....juerg |
Without any personal experience on your particular model, I would say
that the only thermally intermittent components I run into on a daily basis in consumer goods ('scopes shouldn't be any different) that gets better as the set gets warmer is electrolytic caps. Get an ESR meter and go around the 'lytics in that area carefully, and you'll find 'em. "Juerg" wrote in message ps.com... Hi, I could use some help fixing a TDS640 digital Tek scope. The baselines of all four channels show an offset'ed sawtooth (different amplitudes and offsets for each channel) with a period of 4.5ms rather than a flat 0. When applying an input signal, it gets overlaid over the sawtooth but otherwise looks fine. I was following Teks troubleshooting guide and located the problem to the acquisition board. The input attenuator is believed to be OK (input to the AD converters look fine). All the low voltages (+/-15V, +/-5V) are OK and within spec (checked with another scope, no ripples). What I noticed is that over time when the scope warms up the amplitude of the sawtooth decreases until it becomes zero, but there's still an offset that won't go away. I was using some freeze spray to try to isolate the problem further and ended up pin-pointing it to the section around the AD converters. When I cold spray that area, the sawtooth comes back and disappears again after a while. There's a bunch of OpAmps, resisors, caps and diodes in that area. Any ideas what to look at first? Thanks in advance for any suggestions ...juerg |
Juerg wrote:
Hi, I could use some help fixing a TDS640 digital Tek scope. The baselines of all four channels show an offset'ed sawtooth (different amplitudes and offsets for each channel) with a period of 4.5ms rather than a flat 0. When applying an input signal, it gets overlaid over the sawtooth but otherwise looks fine. I was following Teks troubleshooting guide and located the problem to the acquisition board. The input attenuator is believed to be OK (input to the AD converters look fine). All the low voltages (+/-15V, +/-5V) are OK and within spec (checked with another scope, no ripples). What I noticed is that over time when the scope warms up the amplitude of the sawtooth decreases until it becomes zero, but there's still an offset that won't go away. I was using some freeze spray to try to isolate the problem further and ended up pin-pointing it to the section around the AD converters. When I cold spray that area, the sawtooth comes back and disappears again after a while. There's a bunch of OpAmps, resisors, caps and diodes in that area. Any ideas what to look at first? Thanks in advance for any suggestions ....juerg All my experience is with the TDS540, so my comments may or may not be relevant. All the bias voltages are set up by a single D/A converter that gets multiplexed onto hold caps followed by op-amps. Leakage to any other part of the circuit or a defective op-amp will cause the voltage to decay between refreshes. The resultant triangle will appear on the acquired signal. I'd compare the triangle on the waveform to the refresh rate of the D/A hold circuits. Another simple thing to do is dig out your data sheets and check the power supply voltages on all the op-amps and multiplexer chips. I had several +15V supply pins to op-amps go open. ON the 540 series, a major cause of this leakage was failed electrolytic caps that leaked caustic goop onto the board. It can be too little to see and still cause problems. Another problem I've seen is corrosion between op-amp and multiplexer pins facilitated by this electrolyte. Sometimes it gets down tiny blind via holes and eats out the via. You can't get to the other side to test it. Don't know if the 640 series is afflicted with leaky caps. mike -- Return address is VALID but some sites block emails with links. Delete this sig when replying. .. Wanted, PCMCIA SCSI Card for HP m820 CDRW. FS 500MHz Tek DSOscilloscope TDS540 Make Offer Wanted, 12.1" LCD for Gateway Solo 5300. Samsung LT121SU-121 Bunch of stuff For Sale and Wanted at the link below. MAKE THE OBVIOUS CHANGES TO THE LINK htremovethistp://www.geocities.com/SiliconValley/Monitor/4710/ |
Hi Mike,
Thanks for the tips, I'll check them out. You don't happen to have schematics for the TD540, do you? That might help me figuring out the 640. ....juerg mike wrote: Juerg wrote: Hi, I could use some help fixing a TDS640 digital Tek scope. The baselines of all four channels show an offset'ed sawtooth (different amplitudes and offsets for each channel) with a period of 4.5ms rather than a flat 0. When applying an input signal, it gets overlaid over the sawtooth but otherwise looks fine. I was following Teks troubleshooting guide and located the problem to the acquisition board. The input attenuator is believed to be OK (input to the AD converters look fine). All the low voltages (+/-15V, +/-5V) are OK and within spec (checked with another scope, no ripples). What I noticed is that over time when the scope warms up the amplitude of the sawtooth decreases until it becomes zero, but there's still an offset that won't go away. I was using some freeze spray to try to isolate the problem further and ended up pin-pointing it to the section around the AD converters. When I cold spray that area, the sawtooth comes back and disappears again after a while. There's a bunch of OpAmps, resisors, caps and diodes in that area. Any ideas what to look at first? Thanks in advance for any suggestions ....juerg All my experience is with the TDS540, so my comments may or may not be relevant. All the bias voltages are set up by a single D/A converter that gets multiplexed onto hold caps followed by op-amps. Leakage to any other part of the circuit or a defective op-amp will cause the voltage to decay between refreshes. The resultant triangle will appear on the acquired signal. I'd compare the triangle on the waveform to the refresh rate of the D/A hold circuits. Another simple thing to do is dig out your data sheets and check the power supply voltages on all the op-amps and multiplexer chips. I had several +15V supply pins to op-amps go open. ON the 540 series, a major cause of this leakage was failed electrolytic caps that leaked caustic goop onto the board. It can be too little to see and still cause problems. Another problem I've seen is corrosion between op-amp and multiplexer pins facilitated by this electrolyte. Sometimes it gets down tiny blind via holes and eats out the via. You can't get to the other side to test it. Don't know if the 640 series is afflicted with leaky caps. mike -- Return address is VALID but some sites block emails with links. Delete this sig when replying. . Wanted, PCMCIA SCSI Card for HP m820 CDRW. FS 500MHz Tek DSOscilloscope TDS540 Make Offer Wanted, 12.1" LCD for Gateway Solo 5300. Samsung LT121SU-121 Bunch of stuff For Sale and Wanted at the link below. MAKE THE OBVIOUS CHANGES TO THE LINK htremovethistp://www.geocities.com/SiliconValley/Monitor/4710/ |
Juerg wrote:
Hi Mike, Thanks for the tips, I'll check them out. You don't happen to have schematics for the TD540, do you? That might help me figuring out the 640. TDS schematics are unpublished and likely not found anywhere. Anybody who had them would be sitting on a repair bonanza and unlikely to give them up. mike ....juerg mike wrote: Juerg wrote: Hi, I could use some help fixing a TDS640 digital Tek scope. The baselines of all four channels show an offset'ed sawtooth (different amplitudes and offsets for each channel) with a period of 4.5ms rather than a flat 0. When applying an input signal, it gets overlaid over the sawtooth but otherwise looks fine. I was following Teks troubleshooting guide and located the problem to the acquisition board. The input attenuator is believed to be OK (input to the AD converters look fine). All the low voltages (+/-15V, +/-5V) are OK and within spec (checked with another scope, no ripples). What I noticed is that over time when the scope warms up the amplitude of the sawtooth decreases until it becomes zero, but there's still an offset that won't go away. I was using some freeze spray to try to isolate the problem further and ended up pin-pointing it to the section around the AD converters. When I cold spray that area, the sawtooth comes back and disappears again after a while. There's a bunch of OpAmps, resisors, caps and diodes in that area. Any ideas what to look at first? Thanks in advance for any suggestions ....juerg All my experience is with the TDS540, so my comments may or may not be relevant. All the bias voltages are set up by a single D/A converter that gets multiplexed onto hold caps followed by op-amps. Leakage to any other part of the circuit or a defective op-amp will cause the voltage to decay between refreshes. The resultant triangle will appear on the acquired signal. I'd compare the triangle on the waveform to the refresh rate of the D/A hold circuits. Another simple thing to do is dig out your data sheets and check the power supply voltages on all the op-amps and multiplexer chips. I had several +15V supply pins to op-amps go open. ON the 540 series, a major cause of this leakage was failed electrolytic caps that leaked caustic goop onto the board. It can be too little to see and still cause problems. Another problem I've seen is corrosion between op-amp and multiplexer pins facilitated by this electrolyte. Sometimes it gets down tiny blind via holes and eats out the via. You can't get to the other side to test it. Don't know if the 640 series is afflicted with leaky caps. mike -- Return address is VALID but some sites block emails with links. Delete this sig when replying. . Wanted, PCMCIA SCSI Card for HP m820 CDRW. FS 500MHz Tek DSOscilloscope TDS540 Make Offer Wanted, 12.1" LCD for Gateway Solo 5300. Samsung LT121SU-121 Bunch of stuff For Sale and Wanted at the link below. MAKE THE OBVIOUS CHANGES TO THE LINK htremovethistp://www.geocities.com/SiliconValley/Monitor/4710/ -- Return address is VALID but some sites block emails with links. Delete this sig when replying. .. Wanted, PCMCIA SCSI Card for HP m820 CDRW. FS 500MHz Tek DSOscilloscope TDS540 Make Offer Wanted, 12.1" LCD for Gateway Solo 5300. Samsung LT121SU-121 Bunch of stuff For Sale and Wanted at the link below. MAKE THE OBVIOUS CHANGES TO THE LINK htremovethistp://www.geocities.com/SiliconValley/Monitor/4710/ |
I have to agree with mike: The service manuals which can be downloaded
from the usual places (www.tektronix.com, www.teknetelectronics.com) have NO detailed schematics. The go down to moduel level only. So, I'd recommend you try to inspect the board and when you are nesr a D/A converter (AFAIK they used devices from AnaloDevices, marked AD) then try to find the cps nearby. To Juerg: I know this is not a perfect way to diagnose. But what options do you have? The instrument is out of long term support from Tektronix. No schematic avail for the public. I guess then you have to do it the hard way. When you are located in Europe, I might be able to help via phone. hth, Andreas |
That's what I figured from searching the web...
Anyways, I took a closer look today and found the D/A converter (AD667) for the bias voltages and a bunch (5) of analog 8:1 multiplexers (don't remember the part number)... And yes, as Mike suspected, the frequency of the sawtooth is the same as the refresh rate of the D/A. The sawtooth is also visible on the output of a couple of the multiplexers as well as on the input and output of a couple of the OpAmps (TL074C). But I haven't found any holding caps yet, that also show the sawtooth across them. I'll keep hunting around. Maybe I just replace all the electrolytic caps in that area... Thanks for your help, guys ....juerg mike wrote: Juerg wrote: Hi Mike, Thanks for the tips, I'll check them out. You don't happen to have schematics for the TD540, do you? That might help me figuring out the 640. TDS schematics are unpublished and likely not found anywhere. Anybody who had them would be sitting on a repair bonanza and unlikely to give them up. mike ....juerg mike wrote: Juerg wrote: Hi, I could use some help fixing a TDS640 digital Tek scope. The baselines of all four channels show an offset'ed sawtooth (different amplitudes and offsets for each channel) with a period of 4.5ms rather than a flat 0. When applying an input signal, it gets overlaid over the sawtooth but otherwise looks fine. I was following Teks troubleshooting guide and located the problem to the acquisition board. The input attenuator is believed to be OK (input to the AD converters look fine). All the low voltages (+/-15V, +/-5V) are OK and within spec (checked with another scope, no ripples). What I noticed is that over time when the scope warms up the amplitude of the sawtooth decreases until it becomes zero, but there's still an offset that won't go away. I was using some freeze spray to try to isolate the problem further and ended up pin-pointing it to the section around the AD converters. When I cold spray that area, the sawtooth comes back and disappears again after a while. There's a bunch of OpAmps, resisors, caps and diodes in that area. Any ideas what to look at first? Thanks in advance for any suggestions ....juerg All my experience is with the TDS540, so my comments may or may not be relevant. All the bias voltages are set up by a single D/A converter that gets multiplexed onto hold caps followed by op-amps. Leakage to any other part of the circuit or a defective op-amp will cause the voltage to decay between refreshes. The resultant triangle will appear on the acquired signal. I'd compare the triangle on the waveform to the refresh rate of the D/A hold circuits. Another simple thing to do is dig out your data sheets and check the power supply voltages on all the op-amps and multiplexer chips. I had several +15V supply pins to op-amps go open. ON the 540 series, a major cause of this leakage was failed electrolytic caps that leaked caustic goop onto the board. It can be too little to see and still cause problems. Another problem I've seen is corrosion between op-amp and multiplexer pins facilitated by this electrolyte. Sometimes it gets down tiny blind via holes and eats out the via. You can't get to the other side to test it. Don't know if the 640 series is afflicted with leaky caps. mike -- Return address is VALID but some sites block emails with links. Delete this sig when replying. . Wanted, PCMCIA SCSI Card for HP m820 CDRW. FS 500MHz Tek DSOscilloscope TDS540 Make Offer Wanted, 12.1" LCD for Gateway Solo 5300. Samsung LT121SU-121 Bunch of stuff For Sale and Wanted at the link below. MAKE THE OBVIOUS CHANGES TO THE LINK htremovethistp://www.geocities.com/SiliconValley/Monitor/4710/ -- Return address is VALID but some sites block emails with links. Delete this sig when replying. . Wanted, PCMCIA SCSI Card for HP m820 CDRW. FS 500MHz Tek DSOscilloscope TDS540 Make Offer Wanted, 12.1" LCD for Gateway Solo 5300. Samsung LT121SU-121 Bunch of stuff For Sale and Wanted at the link below. MAKE THE OBVIOUS CHANGES TO THE LINK htremovethistp://www.geocities.com/SiliconValley/Monitor/4710/ |
Juerg wrote:
That's what I figured from searching the web... Anyways, I took a closer look today and found the D/A converter (AD667) for the bias voltages and a bunch (5) of analog 8:1 multiplexers (don't remember the part number)... And yes, as Mike suspected, the frequency of the sawtooth is the same as the refresh rate of the D/A. The sawtooth is also visible on the output of a couple of the multiplexers as well as on the input and output of a couple of the OpAmps (TL074C). But I haven't found any holding caps yet, that also show the sawtooth across them. I'll keep hunting around. Maybe I just replace all the electrolytic caps in that area... Thanks for your help, guys ....juerg Look for quad op-amps. TL074 in the 540. They're configured as simple voltage followers. The holding caps are typically 0.1uf SMT on the input. Sometimes there's a series resistor. The caps probably aren't bad, unless they got cracked. There's a lot of stuff on the backside of the board. Make sure to check for VCC on the op-amps. Replacing the electrolytics is good, but won't fix it. If they leaked, the electrolyte is already on the board. Do the math. It takes very little board leakage to turn your reference voltages into triangles. You have to CLEAN the board. First one I fixed, symptoms got better every time I cleaned it. About the fourth time, it started working. Clean the whole board, both sides. I found electrolyte droplets far from any caps. If you try to spot-clean you'll just move the gunk around. Use something like Simple Green and a stiff toothbrush. Clean the CRAP out of it. You have to get the invisible coating out from under the chips. I used an Xacto knife to scrape gunk from between IC pins. Sprayed cleaner HARD in attempt to force it under chips. Cleaned it again with 99% Isopropyl Alcohol. Dried with a hair dryer then let it sit overnight on the heater vent. There's a reason for the two cleanings. Some stuff dissolves in the cleaner that won't in alcohol and vice versa. Took FOUR cycles of this before I got it all working. At the start, I was very timid about scrubbing the board. By the end, my scrubbing was brutal. Tape up any switches on the board and try to keep the cleaners out of 'em. Alcohol probably won't hurt the switches, but all the gunk dissolved in it might. Are we having fun yet? mike OH, about replacing caps...I have never seen your scope, this relates to the TDS540. YMMV. If one is leaking, they probably all are. Change them all. There's over a hundred on a 540. Don't forget the front panel board. You don't want that gunk getting into the encoders. If they're the round silver caps like mine, the connections will be corroded. Go over them one connection at a time to burn off the gunk and get down to clean solder. Then use TWO soldering irons. Apply the heat and twist slightly with the irons to rotate the caps. When it gets hot enough, they spin right off. This technique applies shear force to the pad and is much less likely to lift the pad. Clean it while the caps are off. Mine had two values of caps. I did some voltage checks and decided that 47uF 25V caps would work everywhere. Again, this was a 540 YMMV. Good news was that I had a bunch of 'em. First board, I put a .1uF 1206 smt then put the 47uF leaded cap on top of that. By the fourth board, I left off the .1s. Didn't seem to matter. If you lift a pad, the .1 gives you something stable to solder to. Start on caps where you can get at the trace on both ends. If you lift a pad, you'll be able to fix it. Save the blind ones till you've done a few dozen. Now, we're having fun... -- Return address is VALID but some sites block emails with links. Delete this sig when replying. .. Wanted, PCMCIA SCSI Card for HP m820 CDRW. FS 500MHz Tek DSOscilloscope TDS540 Make Offer Wanted, 12.1" LCD for Gateway Solo 5300. Samsung LT121SU-121 Bunch of stuff For Sale and Wanted at the link below. MAKE THE OBVIOUS CHANGES TO THE LINK htremovethistp://www.geocities.com/SiliconValley/Monitor/4710/ |
mike wrote in :
Juerg wrote: That's what I figured from searching the web... Anyways, I took a closer look today and found the D/A converter (AD667) for the bias voltages and a bunch (5) of analog 8:1 multiplexers (don't remember the part number)... And yes, as Mike suspected, the frequency of the sawtooth is the same as the refresh rate of the D/A. The sawtooth is also visible on the output of a couple of the multiplexers as well as on the input and output of a couple of the OpAmps (TL074C). But I haven't found any holding caps yet, that also show the sawtooth across them. I'll keep hunting around. Maybe I just replace all the electrolytic caps in that area... Thanks for your help, guys ....juerg Look for quad op-amps. TL074 in the 540. They're configured as simple voltage followers. The holding caps are typically 0.1uf SMT on the input. Sometimes there's a series resistor. The caps probably aren't bad, unless they got cracked. There's a lot of stuff on the backside of the board. Make sure to check for VCC on the op-amps. Replacing the electrolytics is good, but won't fix it. If they leaked, the electrolyte is already on the board. Do the math. It takes very little board leakage to turn your reference voltages into triangles. You have to CLEAN the board. First one I fixed, symptoms got better every time I cleaned it. About the fourth time, it started working. Clean the whole board, both sides. I found electrolyte droplets far from any caps. If you try to spot-clean you'll just move the gunk around. Use something like Simple Green and a stiff toothbrush. Clean the CRAP out of it. You have to get the invisible coating out from under the chips. I used an Xacto knife to scrape gunk from between IC pins. Sprayed cleaner HARD in attempt to force it under chips. Cleaned it again with 99% Isopropyl Alcohol. Dried with a hair dryer then let it sit overnight on the heater vent. There's a reason for the two cleanings. Some stuff dissolves in the cleaner that won't in alcohol and vice versa. Took FOUR cycles of this before I got it all working. At the start, I was very timid about scrubbing the board. By the end, my scrubbing was brutal. Tape up any switches on the board and try to keep the cleaners out of 'em. Alcohol probably won't hurt the switches, but all the gunk dissolved in it might. Are we having fun yet? mike OH, about replacing caps...I have never seen your scope, this relates to the TDS540. YMMV. If one is leaking, they probably all are. Change them all. There's over a hundred on a 540. Don't forget the front panel board. You don't want that gunk getting into the encoders. If they're the round silver caps like mine, the connections will be corroded. Go over them one connection at a time to burn off the gunk and get down to clean solder. Then use TWO soldering irons. Apply the heat and twist slightly with the irons to rotate the caps. When it gets hot enough, they spin right off. This technique applies shear force to the pad and is much less likely to lift the pad. Clean it while the caps are off. Mine had two values of caps. I did some voltage checks and decided that 47uF 25V caps would work everywhere. Again, this was a 540 YMMV. Good news was that I had a bunch of 'em. First board, I put a .1uF 1206 smt then put the 47uF leaded cap on top of that. By the fourth board, I left off the .1s. Didn't seem to matter. If you lift a pad, the .1 gives you something stable to solder to. Start on caps where you can get at the trace on both ends. If you lift a pad, you'll be able to fix it. Save the blind ones till you've done a few dozen. Now, we're having fun... If you really want to clean it,run it through your dishwasher(alone,no dishes!) using Calgonite,then dry thoroughly,repeat;thoroughly. We used to run 2236 DMM boards thru a DW to cure a Hi-Z leakage problem. Then 3 days in the drying oven. -- Jim Yanik jyanik at kua.net |
If you really want to clean it,run it through your dishwasher(alone,no dishes!) using Calgonite,then dry thoroughly,repeat;thoroughly. We used to run 2236 DMM boards thru a DW to cure a Hi-Z leakage problem. Then 3 days in the drying oven. I pretty regularly run circuit boards and even entire open-frame video monitors through the dishwasher, people always think I'm crazy but it works so well, the stuff comes out looking like new and is far easier to work on not to mention it runs cooler when I'm done. |
Jim Yanik wrote:
If you really want to clean it,run it through your dishwasher(alone,no dishes!) using Calgonite,then dry thoroughly,repeat;thoroughly. We used to run 2236 DMM boards thru a DW to cure a Hi-Z leakage problem. Then 3 days in the drying oven. -- Jim Yanik Jim, have you seen the small powder coat oven at Harbor Freight? The temperature range is too high, but it looks like a small environmental chamber. For the $399 price it wouldn't be hard to change or modify the controller to use for circuit boards, if needed. http://www.harborfreight.com/cpi/cta...emnumber=46300 -- ? Michael A. Terrell Central Florida |
"Michael A. Terrell" wrote in
: Jim Yanik wrote: If you really want to clean it,run it through your dishwasher(alone,no dishes!) using Calgonite,then dry thoroughly,repeat;thoroughly. We used to run 2236 DMM boards thru a DW to cure a Hi-Z leakage problem. Then 3 days in the drying oven. -- Jim Yanik Jim, have you seen the small powder coat oven at Harbor Freight? The temperature range is too high, but it looks like a small environmental chamber. For the $399 price it wouldn't be hard to change or modify the controller to use for circuit boards, if needed. It is only large enough for boards or small instruments.But it would be a nice oven for a small business. http://www.harborfreight.com/cpi/cta...emnumber=46300 You could make a large plywood box,put a bank or two of 100W light bulbs in it,and a couple of fans drawing air out of the box,for a lot less,IMO. And size it large enough to fit a 7000 series scope mainframe. Gets too hot inside,unscrew a light bulb or two. Not fancy,but workable. -- Jim Yanik jyanik at kua.net |
Jim Yanik wrote:
You could make a large plywood box,put a bank or two of 100W light bulbs in it,and a couple of fans drawing air out of the box,for a lot less,IMO. And size it large enough to fit a 7000 series scope mainframe. Gets too hot inside,unscrew a light bulb or two. Not fancy,but workable. -- Jim Yanik jyanik at kua.net Microdyne used 3 feet high by three feet deep by 8 feet wide plywood boxes with 1500 watt heaters and an external thermostat mounted near the exhaust. They worked, but we were required to add fire extinguishing systems to the cabinets by the fire department. A radio caught fire during burn-in and scorched the plywood. It was a sad sight. A new $20,000 radio with several circuit boards burnt to bits of burnt fiberglass and loose pieces of copper clad. -- ? Michael A. Terrell Central Florida |
On Wed, 09 Mar 2005 07:47:37 GMT "Michael A. Terrell"
wrote: Jim, have you seen the small powder coat oven at Harbor Freight? The temperature range is too high, but it looks like a small environmental chamber. For the $399 price it wouldn't be hard to change or modify the controller to use for circuit boards, if needed. You don't have to get fancy if you just need to do a single part (or a few parts.) I've done wonders with a steel wastebasket and a small light bulb. You adjust the temp by trying different size bulbs, and you close off the wastebasket opening with alum foil, just to limit air circulation, which would reduce the temp. Make sure the bulb is placed so that it can't come in direct contact with the parts you're drying out. It doesn't have to be hot, just warm enough to make the vapor pressure of the air low (ie, reduce the relative humidity.) 125-150F is plenty warm. - ----------------------------------------------- Jim Adney Madison, WI 53711 USA ----------------------------------------------- |
Jim Adney wrote in
: On Wed, 09 Mar 2005 07:47:37 GMT "Michael A. Terrell" wrote: Jim, have you seen the small powder coat oven at Harbor Freight? The temperature range is too high, but it looks like a small environmental chamber. For the $399 price it wouldn't be hard to change or modify the controller to use for circuit boards, if needed. I think Mike is considering commercial uses,not home/hobby applications. For that(commercial apps) it probably would be a good idea. I'd want something larger,that I could put a whole instrument in after a wash. You don't have to get fancy if you just need to do a single part (or a few parts.) I've done wonders with a steel wastebasket and a small light bulb. You adjust the temp by trying different size bulbs, and you close off the wastebasket opening with alum foil, just to limit air circulation, which would reduce the temp. Make sure the bulb is placed so that it can't come in direct contact with the parts you're drying out. It doesn't have to be hot, just warm enough to make the vapor pressure of the air low (ie, reduce the relative humidity.) 125-150F is plenty warm. Actually,you do NOT want it any hotter than about 150degF, plastic parts like cam drums begin to be affected. And the muffin fans providing a slight negative pressure helps draw out the moisture,increasing evaporation. (BTW,I've also used the lightbulb/enclosure trick for curing epoxy.) -- Jim Yanik jyanik at kua.net |
Jim Adney wrote:
On Wed, 09 Mar 2005 07:47:37 GMT "Michael A. Terrell" wrote: Jim, have you seen the small powder coat oven at Harbor Freight? The temperature range is too high, but it looks like a small environmental chamber. For the $399 price it wouldn't be hard to change or modify the controller to use for circuit boards, if needed. You don't have to get fancy if you just need to do a single part (or a few parts.) I've done wonders with a steel wastebasket and a small light bulb. You adjust the temp by trying different size bulbs, and you close off the wastebasket opening with alum foil, just to limit air circulation, which would reduce the temp. Make sure the bulb is placed so that it can't come in direct contact with the parts you're drying out. It doesn't have to be hot, just warm enough to make the vapor pressure of the air low (ie, reduce the relative humidity.) 125-150F is plenty warm. Jim Adney I was thinking about baking large SMD ICs before soldering them in a reflow oven, or with a hot air rework station. If you don't, that can cause cracks in the moulded body and destroy the chip. I saw a couple hundred $80 FIR chips destroyed by assembly because they ignored the warning when the set up production of a new circuit board. We also baked chips used by rework, then put a descant pack with the chip in a modified carrier and heat sealed it in a moisture proof bag. A few boards were built with a large chip turned 90 degrees (Or more) The boards were backed prior to rework to lift and rotate the part. Repair of some equipment will move in this direction before you know it. -- ? Michael A. Terrell Central Florida |
Jim Yanik wrote:
Jim Adney wrote in : On Wed, 09 Mar 2005 07:47:37 GMT "Michael A. Terrell" wrote: Jim, have you seen the small powder coat oven at Harbor Freight? The temperature range is too high, but it looks like a small environmental chamber. For the $399 price it wouldn't be hard to change or modify the controller to use for circuit boards, if needed. I think Mike is considering commercial uses,not home/hobby applications. For that(commercial apps) it probably would be a good idea. I'd want something larger,that I could put a whole instrument in after a wash. You don't have to get fancy if you just need to do a single part (or a few parts.) I've done wonders with a steel wastebasket and a small light bulb. You adjust the temp by trying different size bulbs, and you close off the wastebasket opening with alum foil, just to limit air circulation, which would reduce the temp. Make sure the bulb is placed so that it can't come in direct contact with the parts you're drying out. It doesn't have to be hot, just warm enough to make the vapor pressure of the air low (ie, reduce the relative humidity.) 125-150F is plenty warm. Actually,you do NOT want it any hotter than about 150degF, plastic parts like cam drums begin to be affected. And the muffin fans providing a slight negative pressure helps draw out the moisture,increasing evaporation. (BTW,I've also used the lightbulb/enclosure trick for curing epoxy.) Jim Yanik Yes, I was talking about professional use. I've looked at the oven in the Ocala store. It would handle most circuit boards and small equipment. I didn't have a tape measure with me to get the inside dimensions. I would like one for my small shop to do multi layer circuit board repairs, but those days are likely behind me. -- ? Michael A. Terrell Central Florida |
OK, here's an update on my progress:
I cleaned the acquisition board using all methods mentioned in the thread (toothbrush with simple green and isopropyl alcohol followed by a dishwasher cycle). I reassembled the scope and - oh wonder - the sawtooth/triangle is gone but all the channels still have different offsets (even when the input is grounded). Could that still be an after effect of leaky caps? I'll start looking at the bias voltage generation/convertion to figure where the error gets introduced. NOW I'm having fun :-) ....juerg mike wrote: Juerg wrote: That's what I figured from searching the web... Anyways, I took a closer look today and found the D/A converter (AD667) for the bias voltages and a bunch (5) of analog 8:1 multiplexers (don't remember the part number)... And yes, as Mike suspected, the frequency of the sawtooth is the same as the refresh rate of the D/A. The sawtooth is also visible on the output of a couple of the multiplexers as well as on the input and output of a couple of the OpAmps (TL074C). But I haven't found any holding caps yet, that also show the sawtooth across them. I'll keep hunting around. Maybe I just replace all the electrolytic caps in that area... Thanks for your help, guys ....juerg Look for quad op-amps. TL074 in the 540. They're configured as simple voltage followers. The holding caps are typically 0.1uf SMT on the input. Sometimes there's a series resistor. The caps probably aren't bad, unless they got cracked. There's a lot of stuff on the backside of the board. Make sure to check for VCC on the op-amps. Replacing the electrolytics is good, but won't fix it. If they leaked, the electrolyte is already on the board. Do the math. It takes very little board leakage to turn your reference voltages into triangles. You have to CLEAN the board. First one I fixed, symptoms got better every time I cleaned it. About the fourth time, it started working. Clean the whole board, both sides. I found electrolyte droplets far from any caps. If you try to spot-clean you'll just move the gunk around. Use something like Simple Green and a stiff toothbrush. Clean the CRAP out of it. You have to get the invisible coating out from under the chips. I used an Xacto knife to scrape gunk from between IC pins. Sprayed cleaner HARD in attempt to force it under chips. Cleaned it again with 99% Isopropyl Alcohol. Dried with a hair dryer then let it sit overnight on the heater vent. There's a reason for the two cleanings. Some stuff dissolves in the cleaner that won't in alcohol and vice versa. Took FOUR cycles of this before I got it all working. At the start, I was very timid about scrubbing the board. By the end, my scrubbing was brutal. Tape up any switches on the board and try to keep the cleaners out of 'em. Alcohol probably won't hurt the switches, but all the gunk dissolved in it might. Are we having fun yet? mike OH, about replacing caps...I have never seen your scope, this relates to the TDS540. YMMV. If one is leaking, they probably all are. Change them all. There's over a hundred on a 540. Don't forget the front panel board. You don't want that gunk getting into the encoders. If they're the round silver caps like mine, the connections will be corroded. Go over them one connection at a time to burn off the gunk and get down to clean solder. Then use TWO soldering irons. Apply the heat and twist slightly with the irons to rotate the caps. When it gets hot enough, they spin right off. This technique applies shear force to the pad and is much less likely to lift the pad. Clean it while the caps are off. Mine had two values of caps. I did some voltage checks and decided that 47uF 25V caps would work everywhere. Again, this was a 540 YMMV. Good news was that I had a bunch of 'em. First board, I put a .1uF 1206 smt then put the 47uF leaded cap on top of that. By the fourth board, I left off the .1s. Didn't seem to matter. If you lift a pad, the .1 gives you something stable to solder to. Start on caps where you can get at the trace on both ends. If you lift a pad, you'll be able to fix it. Save the blind ones till you've done a few dozen. Now, we're having fun... -- Return address is VALID but some sites block emails with links. Delete this sig when replying. . Wanted, PCMCIA SCSI Card for HP m820 CDRW. FS 500MHz Tek DSOscilloscope TDS540 Make Offer Wanted, 12.1" LCD for Gateway Solo 5300. Samsung LT121SU-121 Bunch of stuff For Sale and Wanted at the link below. MAKE THE OBVIOUS CHANGES TO THE LINK htremovethistp://www.geocities.com/SiliconValley/Monitor/4710/ |
I managed to get rid of the offsets by running the internal signal path
calibration. The scope is now fully functional and works like a charm :-) thanks for all the help! |
Did you try running signal path compensation.
It usually takes care of signal offsets. Go to Utility-Cal, read the screen and run SPC. If everything is fine, you'll have no DC offset any more. --Artem Juerg wrote: OK, here's an update on my progress: I cleaned the acquisition board using all methods mentioned in the thread (toothbrush with simple green and isopropyl alcohol followed by a dishwasher cycle). I reassembled the scope and - oh wonder - the sawtooth/triangle is gone but all the channels still have different offsets (even when the input is grounded). Could that still be an after effect of leaky caps? I'll start looking at the bias voltage generation/convertion to figure where the error gets introduced. NOW I'm having fun :-) ...juerg mike wrote: Juerg wrote: That's what I figured from searching the web... Anyways, I took a closer look today and found the D/A converter (AD667) for the bias voltages and a bunch (5) of analog 8:1 multiplexers (don't remember the part number)... And yes, as Mike suspected, the frequency of the sawtooth is the same as the refresh rate of the D/A. The sawtooth is also visible on the output of a couple of the multiplexers as well as on the input and output of a couple of the OpAmps (TL074C). But I haven't found any holding caps yet, that also show the sawtooth across them. I'll keep hunting around. Maybe I just replace all the electrolytic caps in that area... Thanks for your help, guys ....juerg Look for quad op-amps. TL074 in the 540. They're configured as simple voltage followers. The holding caps are typically 0.1uf SMT on the input. Sometimes there's a series resistor. The caps probably aren't bad, unless they got cracked. There's a lot of stuff on the backside of the board. Make sure to check for VCC on the op-amps. Replacing the electrolytics is good, but won't fix it. If they leaked, the electrolyte is already on the board. Do the math. It takes very little board leakage to turn your reference voltages into triangles. You have to CLEAN the board. First one I fixed, symptoms got better every time I cleaned it. About the fourth time, it started working. Clean the whole board, both sides. I found electrolyte droplets far from any caps. If you try to spot-clean you'll just move the gunk around. Use something like Simple Green and a stiff toothbrush. Clean the CRAP out of it. You have to get the invisible coating out from under the chips. I used an Xacto knife to scrape gunk from between IC pins. Sprayed cleaner HARD in attempt to force it under chips. Cleaned it again with 99% Isopropyl Alcohol. Dried with a hair dryer then let it sit overnight on the heater vent. There's a reason for the two cleanings. Some stuff dissolves in the cleaner that won't in alcohol and vice versa. Took FOUR cycles of this before I got it all working. At the start, I was very timid about scrubbing the board. By the end, my scrubbing was brutal. Tape up any switches on the board and try to keep the cleaners out of 'em. Alcohol probably won't hurt the switches, but all the gunk dissolved in it might. Are we having fun yet? mike OH, about replacing caps...I have never seen your scope, this relates to the TDS540. YMMV. If one is leaking, they probably all are. Change them all. There's over a hundred on a 540. Don't forget the front panel board. You don't want that gunk getting into the encoders. If they're the round silver caps like mine, the connections will be corroded. Go over them one connection at a time to burn off the gunk and get down to clean solder. Then use TWO soldering irons. Apply the heat and twist slightly with the irons to rotate the caps. When it gets hot enough, they spin right off. This technique applies shear force to the pad and is much less likely to lift the pad. Clean it while the caps are off. Mine had two values of caps. I did some voltage checks and decided that 47uF 25V caps would work everywhere. Again, this was a 540 YMMV. Good news was that I had a bunch of 'em. First board, I put a .1uF 1206 smt then put the 47uF leaded cap on top of that. By the fourth board, I left off the .1s. Didn't seem to matter. If you lift a pad, the .1 gives you something stable to solder to. Start on caps where you can get at the trace on both ends. If you lift a pad, you'll be able to fix it. Save the blind ones till you've done a few dozen. Now, we're having fun... -- Return address is VALID but some sites block emails with links. Delete this sig when replying. . Wanted, PCMCIA SCSI Card for HP m820 CDRW. FS 500MHz Tek DSOscilloscope TDS540 Make Offer Wanted, 12.1" LCD for Gateway Solo 5300. Samsung LT121SU-121 Bunch of stuff For Sale and Wanted at the link below. MAKE THE OBVIOUS CHANGES TO THE LINK htremovethistp://www.geocities.com/SiliconValley/Monitor/4710/ |
I should start reading messages newest first... I've just fired a
message suggesting exactly this. Congratulations, anyways. --Artem Juerg wrote: I managed to get rid of the offsets by running the internal signal path calibration. The scope is now fully functional and works like a charm :-) thanks for all the help! |
All times are GMT +1. The time now is 06:59 PM. |