Tektronix 2467B Oscilloscope Error Codes
Hi, I am getting these error codes on startup on my Tektronix 2467B Scope. I get "CT Test 84 Fail 40" I don't think it is just a calibration error as I have been lead to beleive. After this you can jit A/B Trigger to Exit and you get Two rows of Dots on the screen. Any help would be appreciated I love this scope and want to fix it properly.....................Bill................. .
-- Posted by Mimo Usenet Browser v0.2.4 http://www.goldenfrog.com/mimo/post |
Tektronix 2467B Oscilloscope Error Codes
On Thu, 15 Dec 2011 17:31:14 -0600 "William T Barry"
wrote in Message id: : [reformatted for your reading pleasure] Hi, I am getting these error codes on startup on my Tektronix 2467B Scope. I get "CT Test 84 Fail 40" I don't think it is just a calibration error as I have been lead to beleive. After this you can jit A/B Trigger to Exit and you get Two rows of Dots on the screen. Any help would be appreciated I love this scope and want to fix it properly.....................Bill................. . Hi Bill, It's not the scope that's failing, it's the counter timer option that's failing. You need to find a service manual for that option. If worse comes to worst, it's possible to remove the option. I've done it before a long time ago. IIRC there's some jumpers on the main board that need to be added when the cables connected to the CT are removed. |