#1
![]()
Posted to sci.electronics.repair
|
|||
|
|||
![]()
I have a tek 2215A CRO with following fault..
Left side of trace has a ripple, about 1/2 division vert amplitude and about 25% of horizontal is affected. Both channels the same When vert offset is manually swept, the pattern of ripple changes, still both channels the same Timebase does not change shape or size of ripple ripple is 1/2 or 1 cycle but is like someone has put a magnet on the side of the tube. Changing attenuator makes no difference. Was OK a few months ago when last used If this is a tube fault, I will junk it, what do others think? |
#2
![]()
Posted to sci.electronics.repair
|
|||
|
|||
![]()
On 11/8/2011 6:51 PM, Geoff wrote:
I have a tek 2215A CRO with following fault.. Left side of trace has a ripple, about 1/2 division vert amplitude and about 25% of horizontal is affected. Both channels the same When vert offset is manually swept, the pattern of ripple changes, still both channels the same Timebase does not change shape or size of ripple ripple is 1/2 or 1 cycle but is like someone has put a magnet on the side of the tube. Changing attenuator makes no difference. Was OK a few months ago when last used If this is a tube fault, I will junk it, what do others think? A screenshot would be useful, if it worked ok previously and hasn't been physically dropped/damaged its unlikely the CRT would suddenly develop a distortion, although your description sounds like a duff tube. JC |
#3
![]()
Posted to sci.electronics.repair
|
|||
|
|||
![]()
"Geoff" wrote in message
... I have a tek 2215A CRO with following fault.. Left side of trace has a ripple, about 1/2 division vert amplitude and about 25% of horizontal is affected. Both channels the same When vert offset is manually swept, the pattern of ripple changes, still both channels the same Timebase does not change shape or size of ripple ripple is 1/2 or 1 cycle but is like someone has put a magnet on the side of the tube. Changing attenuator makes no difference. Was OK a few months ago when last used If this is a tube fault, I will junk it, what do others think? Always start with the power supply. Mark Z. |
#4
![]()
Posted to sci.electronics.repair
|
|||
|
|||
![]() "Geoff" Left side of trace has a ripple, about 1/2 division vert amplitude and about 25% of horizontal is affected. Both channels the same When vert offset is manually swept, the pattern of ripple changes, still both channels the same Timebase does not change shape or size of ripple ripple is 1/2 or 1 cycle but is like someone has put a magnet on the side of the tube. ** Could the case be magnetised in one area ? .... Phil |
#5
![]()
Posted to sci.electronics.repair
|
|||
|
|||
![]()
Geoff wrote in message
... I have a tek 2215A CRO with following fault.. Left side of trace has a ripple, about 1/2 division vert amplitude and about 25% of horizontal is affected. Both channels the same When vert offset is manually swept, the pattern of ripple changes, still both channels the same Timebase does not change shape or size of ripple ripple is 1/2 or 1 cycle but is like someone has put a magnet on the side of the tube. Changing attenuator makes no difference. Was OK a few months ago when last used If this is a tube fault, I will junk it, what do others think? I would suspect something in the flyback suppression/blanking area |
#6
![]()
Posted to sci.electronics.repair
|
|||
|
|||
![]()
"Mark Zacharias" wrote in
.com: "Geoff" wrote in message ... I have a tek 2215A CRO with following fault.. Left side of trace has a ripple, about 1/2 division vert amplitude and about 25% of horizontal is affected. Both channels the same When vert offset is manually swept, the pattern of ripple changes, still both channels the same Timebase does not change shape or size of ripple ripple is 1/2 or 1 cycle but is like someone has put a magnet on the side of the tube. some CRTs had tiny magnets (0.1" dia.) glued on at the factory,for beam alignment. one could have come loose. Changing attenuator makes no difference. Was OK a few months ago when last used If this is a tube fault, I will junk it, what do others think? Always start with the power supply. Yes,PS filter caps,and other decoupling caps on the boards are a problem in aging scopes. Mark Z. to check tube geometry,input a 6 div high signal with a fast rise time(like a square wave or time marks),one cycle per division,and turn up intensity to see the fast rise. if the tube is bad,you'll see distorted lines. the distortion remains in the same place as you move trace position controls. IMO,2215-series are good scopes,easy to repair. I have a 2213,one that I built myself from boards discarded by other techs. Scored a second-grade CRT from the head of CRT mfg,bought the little bits I needed to finish the scope for about $20. good analog scopes are hard to find these days. -- Jim Yanik jyanik at localnet dot com |
#7
![]()
Posted to sci.electronics.repair
|
|||
|
|||
![]()
On Tue, 8 Nov 2011 23:51:53 +0000 (UTC), Geoff
wrote: I have a tek 2215A CRO with following fault.. Left side of trace has a ripple, about 1/2 division vert amplitude and about 25% of horizontal is affected. Both channels the same When vert offset is manually swept, the pattern of ripple changes, still both channels the same I don't know this scope but used to work on another manufacturer's models. On one in particular, you could sometimes see this fault - it got worse as the y shift was changed. As I recall, the problem was timebase ripple (after flyback) on the supply to one of the tube accelerating electrodes. The effect was more noticable as the beam was moved (y shift) nearer to the electrode. You would need a second scope to probe the tube pins to confirm this. |
#8
![]()
Posted to sci.electronics.repair
|
|||
|
|||
![]()
"Phil Allison" wrote in
: "Geoff" Left side of trace has a ripple, about 1/2 division vert amplitude and about 25% of horizontal is affected. Both channels the same When vert offset is manually swept, the pattern of ripple changes, still both channels the same Timebase does not change shape or size of ripple ripple is 1/2 or 1 cycle but is like someone has put a magnet on the side of the tube. ** Could the case be magnetised in one area ? ... Phil I took the case off and no difference, your implied suggestion that a change in magnetic field caused this matches what I observe, ie no electronic settings change the waveform, it looks like when someone shoves a magnet near a crt. I have not removed the tube screen but I have shaken it and nothing like a magnet bouncing around can be heard, I may just give up now, it is not a paying job just an old CRO we use occasionally for R&D work, we have newer better ones around including tek analog CRO's. |
#9
![]()
Posted to sci.electronics.repair
|
|||
|
|||
![]() "Geoff" I took the case off and no difference, your implied suggestion that a change in magnetic field caused this matches what I observe, ie no electronic settings change the waveform, it looks like when someone shoves a magnet near a crt. I have not removed the tube screen but I have shaken it and nothing like a magnet bouncing around can be heard, ** The CRT is a *mesh PDA* type - right ? There must be a chance the steel mesh has become magnetised. Suggest you play around with a small magnet near the tube face and see what happens. FYI: There is a fine steel mesh just behind the face of the tube that has final acceleration voltage applied to it - if you de-focus the beam you can usually see the pattern of the mesh projected onto the phosphor. .... Phil |
#10
![]()
Posted to sci.electronics.repair
|
|||
|
|||
![]()
On Nov 10, 1:31*am, "Phil Allison" wrote:
"Geoff" I took the case off and no difference, your implied suggestion that a change in magnetic field caused this matches what I observe, ie no electronic settings change the waveform, it looks like when someone shoves a magnet near a crt. I have not removed the tube screen but I have shaken it and nothing like a magnet bouncing around can be heard, ** The CRT is a **mesh PDA* type *- *right ? There must be a chance the steel mesh has become magnetised. Suggest you play around with a small magnet near the tube face and see what happens. FYI: There is a fine steel mesh just behind the face of the tube that has final acceleration voltage applied to it - *if you de-focus the beam you can usually see the pattern of the mesh projected onto the phosphor. ... *Phil Use a degaussing coil like you use on a color CRT TV set!! |
#11
![]()
Posted to sci.electronics.repair
|
|||
|
|||
![]()
"Phil Allison" wrote in
: "Geoff" I took the case off and no difference, your implied suggestion that a change in magnetic field caused this matches what I observe, ie no electronic settings change the waveform, it looks like when someone shoves a magnet near a crt. I have not removed the tube screen but I have shaken it and nothing like a magnet bouncing around can be heard, ** The CRT is a *mesh PDA* type - right ? There must be a chance the steel mesh has become magnetised. Suggest you play around with a small magnet near the tube face and see what happens. FYI: There is a fine steel mesh just behind the face of the tube that has final acceleration voltage applied to it - if you de-focus the beam you can usually see the pattern of the mesh projected onto the phosphor. ... Phil Links below for pics taken with webcam. Mesh is visible. I have played around with a magnet, can get trace to wobble around as expected but it comes back to the same position once magnet removed. This is true for the unshielded part behind screen and the conical tube shield. I will try a degausser once I find or make one http://i251.photobucket.com/albums/g...01/focus_limit _ccw1.jpg http://i251.photobucket.com/albums/g...01/focus_limit _cw1.jpg http://i251.photobucket.com/albums/g...01/focus_limit _ccw2.jpg http://i251.photobucket.com/albums/g...e101/2215a.jpg http://i251.photobucket.com/albums/g...e101/2215b.jpg |
#12
![]()
Posted to sci.electronics.repair
|
|||
|
|||
![]()
On 11/13/2011 8:13 PM, Geoff wrote:
Links below for pics taken with webcam. Mesh is visible. I have played around with a magnet, can get trace to wobble around as expected but it comes back to the same position once magnet removed. This is true for the unshielded part behind screen and the conical tube shield. I will try a degausser once I find or make one http://i251.photobucket.com/albums/g...01/focus_limit _ccw1.jpg http://i251.photobucket.com/albums/g...01/focus_limit _cw1.jpg http://i251.photobucket.com/albums/g...01/focus_limit _ccw2.jpg http://i251.photobucket.com/albums/g...e101/2215a.jpg http://i251.photobucket.com/albums/g...e101/2215b.jpg If the distortion stays constant with the traces as they move up the screen its not the CRT If the distortion is fixed in that area of the CRT then the CRT is internally damaged. Magnets and deguassers will not cure this. IMO its a broken CRT JC |
Reply |
Thread Tools | Search this Thread |
Display Modes | |
|
|
![]() |
||||
Thread | Forum | |||
Repair Tektronix tds520 | Electronics Repair | |||
tektronix 2212 repair | Electronics Repair | |||
Looking for Service Manual for Tek 2215A | Electronics Repair | |||
Tektronix 2215A Service Manual | Electronics Repair | |||
Tektronix TDS540 repair | Electronics Repair |